Initial Assessment

  • Begin with manual ventilation
  • Assess chest compliance
  • Check for respiratory obstruction

Initial Settings

  • FiO₂: 1.0 initially
  • Vt: 6-8 mL/kg (500 mL adult)
  • RR: 12-15 breaths/min
  • PEEP: 5 cmH₂O
  • I:E ratio: 1:2

Monitoring

  • Check ABG at initiation
  • Repeat ABG after 20 minutes
  • Adjust settings based on results

Troubleshooting (POPES)

  • P: Position of tube
  • O: Obstruction
  • P: Patient issues
  • E: Equipment failure
  • S: Splinting issues

Emergency Response

  • Switch to hand ventilation
  • Systematically check equipment
  • Assess patient condition

Ventilator Setup and Troubleshooting Guide

Comprehensive reference for mechanical ventilation management and problem-solving

Basic Principles

Initially the patient's lungs are ventilated manually, which enables an assessment of chest compliance and assists in the detection of any respiratory obstruction.

Initial Ventilator Settings

Pre-connection Check

Before attaching to the patient, perform a quick ventilator check to ensure proper functioning and prevent complications.

Connect O2 supply via a Schrader valve to the wall supply or cylinder and turn the ventilator on.
For a 'standard' adult set FiO2 to 1.0.
Set the mode to intermittent positive pressure ventilation (IPPV).
Set a tidal volume (VT) of 6 - 8 mL/kg (500 mL for a 'standard' adult) and an appropriate respiratory rate (RR) to achieve a minute volume (MV) of 100 mL/kg (15 breaths/minute for a 'standard' adult).
Set the peak airway pressure (Pmax) to 40 cmH2O.
Set the positive end expiratory pressure (PEEP) to 5 cmH2O.
Set inspiratory:expiratory (I:E) ratio to 1:2.
Connect to a test 'lung' (or use the reservoir bag of an anaesthetic breathing system).
Check that the 'lung' inflates and deflates.
Squeeze the 'lung' to simulate obstruction or increasing airway pressures and check that the P极ax alarm sounds.
Disconnect the circuit and check that the disconnect alarm sounds.
Connect to the patient.
Adjust VT according to clinical chest expansion and peak airway pressures (< 20 cmH2O with normal lungs).

Ventilator Troubleshooting

Immediate Action

It is safest to switch to hand ventilation whilst solving a ventilation problem. If you can adequately hand ventilate the patient then systematically check for ventilatory or circuit issues.

POPES Mnemonic for Ventilation Failure

P: Position of tube

  • Oesophageal or bronchial intubation
  • Tube out of trachea
  • Tube displacement

O: Obstruction

  • Tubing obstruction
  • Tracheal/bronchial obstruction
  • Bronchoconstriction
  • Secretions blocking airway

P: Patient
  • Pneumothorax
  • Pulmonary embolism
  • Pulmonary oedema
  • Atelectasis
  • Bronchospasm

E: Equipment

  • Incorrect kit
  • Connection/equipment failure
  • Inadequate gas flow
  • Power failure
  • Leaks in the system

S: Splinting

  • Abdominal distension including stomach distension
  • Poor paralysis
  • Patient-ventilator asynchrony
  • Inadequate sedation

Common Problems and Solutions

Problem Solution
Not ventilating: no noise or pressure generated
  • Check on/off switch
  • Check O2 source/connections
  • Check power supply
Not ventilating: noise but no pressure generated
  • Check for disconnections
  • Check ventilation valve
  • Check for tracheal cuff leak
  • Check circuit integrity极li>
Not ventilating: noise and pressure generated
  • Check for tubing kinks or obstruction
  • Patient may be fighting the ventilator
  • May have stiff lungs
  • Possible tension pneumothorax
Hypoxaemia
  • Check clinical air entry
  • Increase FiO2
  • Increase PEEP
  • Consider tracheal suction
  • Check tube position
Hypercapnoea
  • Increase MV by increasing RR
  • Increase tidal volume if possible
  • Check for increased dead space
Hypocapnoea
  • Decrease MV by decreasing RR
  • Decrease tidal volume if appropriate
  • Consider patient's metabolic needs

Other Considerations

Tube Position Verification

Obtain chest x-ray to check position of tracheal and nasogastric tubes. Clinical assessment alone is not sufficient to confirm proper placement.

Gastric Decompression

Insertion of a nasogastric or orogastric tube into the stomach enables deflation of a dilated stomach and removal of stomach contents; this helps to reduce the potential for aspiration and may sometimes enable lower ventilation pressures.

Chest Drain Consideration

Consider prophylactic chest drains in chest trauma, particularly if the patient is about to be transported or undergo a lengthy procedure where the chest will not be immediately accessible.

Ongoing Management

Continuous monitoring of ventilator parameters, patient response, and regular arterial blood gas analysis is essential for optimal ventilator management. Adjust settings based on clinical response and laboratory results.